Single-point calibrations of
temperature and rel. humidity
CAL1: With CAL1 (single-point calibration)
you set the offset for Sensor 1 (temperature).
The offset is shown in the lower part of
the display. Maximum offset is ± 10 °C or
± 10 °F.
You can select with
and
with
.
The original works settings can be
achieved by setting the offset to 0.0.
and confirm
Operating manual – English
CAL2: With CAL2 (single-point calibration)
you set the offset for Sensor 2 (rel. humidity).
The offset turns the characteristic line about
the lower equalisation point (11 % r.h.). The
equalisation point must be in the range of
30 % ... 95 % r.h. The offset is shown in the
upper half of the display. Maximum offset
is ± 10 % rH. CAL2 can only be selected in
connection with the unit % r.h.
You can select with
and
with
.
The original works settings are
achieved by setting the offset to 0.0.
and confirm
B - 06